According to the AMS-2750, 3 ways could be used to perform SAT.
1. Conventional SAT, this requirement involves temperature comparison between equipment temperature measurement system and calibration system. For gas furnace,to eliminate the effect of gas flow on temperature sensor, we recommended use Double-element thermocouple(双芯热电偶), of which one can be used as the equipment sensor and the other as test sensor. However, there are additional requirement for the type of the thermocouples used.
2. Alternate SAT. this method involves calibrate the equipment sensor and remain system(the combination of connector, wire, instrument) individually and calculate the SAT error algebraically. This method do not requires a real ramp up on furnace but have special requirement on the replacement of sensors.
3. SAT Waiver. We do not recommend this method. |