热处理论坛

 找回密码
 免费注册

QQ登录

只需一步,快速开始

北京中仪天信科技有限公司 江苏东方航天校准检测有限公司 热处理论坛合作伙伴广告位
搜索
查看: 21979|回复: 3

[Discussion] AMS 2750F About calibration accuracy of sensor

[复制链接]

签到天数: 13 天

[LV.3]偶尔看看II

发表于 2021-3-8 16:09:47 | 显示全部楼层 |阅读模式
北京中仪天信科技有限公司
Hello everybody, i have a question  about calibration accuracy of sensor as below,
According to AMS 2750F table 1,if the calibration accuracy of the type S control sensor is exceed the ±0.6℃ or±0.1%,
shall we scrap the sensor or continue to use the sensor with correction factors?
Hope your answer!

签到天数: 15 天

[LV.4]偶尔看看III

发表于 2021-4-9 13:14:18 来自手机 | 显示全部楼层
You shall use the sensor with correction factor. see table1 note(9).
回复 支持 反对

使用道具 举报

签到天数: 96 天

[LV.6]常住居民II

发表于 2021-4-15 09:32:02 | 显示全部楼层
In my opinion, we have to scrap these sensors becuase they can't mee the requirements.

For table 1 Note 9
I think it is only suitable for sensors type not listed in Table 1 and used above 1316℃ then we can follow up the rules of Note 9.
回复 支持 反对

使用道具 举报

签到天数: 21 天

[LV.4]偶尔看看III

发表于 2021-6-21 11:19:43 来自手机 | 显示全部楼层
See table1 note(9).....
回复 支持 反对

使用道具 举报

您需要登录后才可以回帖 登录 | 免费注册

本版积分规则

QQ|Archiver|手机版|小黑屋|热处理论坛 ( 苏ICP备2021037530号|苏公网安备32059002001695号 )
Copyright © 2005-2023, rclbbs.com 苏州热协网络技术有限公司 版权所有

GMT+8, 2024-3-28 18:22 , Processed in 0.041295 second(s), 20 queries .

Powered by Discuz! X3.4

Copyright © 2001-2021, Tencent Cloud.

快速回复 返回顶部 返回列表